  • Melexis MLX90632 Thermopile

    I'm trying to make heads or tails of the necessary I2C calls to get this to continuously output the raw sensor data. Not sure if there's a general call necessary but also the registers involved seem a bit buried or maybe I'm just not getting it. .PDF attached
    Attached Files

  • #2
    Keith, It looks like you can place the device into CONTINOUS triggering mode but, you will have to set the address each time you retrieve the 6 results.

    The sequence should look like the following:



    Write new_data = 0


    Check when new_data = 1


    Read cycle_pos to get measurement pointer


    If cycle_pos = 1


    Calculate TA and TO based on RAM_4, RAM_5, RAM_6, RAM_9


    If cycle_pos = 2


    Calculate TA and TO based on RAM_7, RAM_8, RAM_6, RAM_9


    Return to top
    Dave Purola,
    N8NTA
    EN82fn

    Comment


    • #3
      Thanks, I'll see if I can make sense of what you're saying by finding those registers. I am also just looking for the raw data and the low byte at that.